| Limit |
|
Testing method | Standard | Frequency range MHz | dB V/m | Remark |
|
|
| Quasi-peak |
|
OATS a or SAC bd | CISPR 16-2-3 | 30 230 |
| 30 | Measurement distance 10 m |
230 300 |
| 37 |
300 | 1 000 | 37 |
|
FAR e | CISPR 16-2-3 | 30 230 | 230 1 000 | 42 to 35 f 42 | Measurement distance 3 m |
|
| 30 | 230 | 30 |
|
TEM-Waveguidec | IEC 61000-4-20 | 230 | 1 000 | 37 | |
NOTE | The lower limit is applicable at the transition frequency. |
a | OATS = open area test site |
b | SAC = semi-anechoic chamber |
c | The TEM-waveguide is limited to devices without cables attached and with a maximum size according to subclause 6.1 of IEC 61000-4-20 (The largest dimension of the enclosure at 1 GHz measuring frequency is one wavelength, 300 mm at 1 GHz) |
d | Measurements may be made at closer distance, down to 3 m. An inverse proportionality factor of 20 dB per decade shall be used to normalize the measured data to the specified distance for determining compliance. |
e | FAR = fully anechoic room. All equipment, including floor-standing equipment, shall be measured within the test volume as described in Figure 6 of CISPR 16-2-3. |
f | Decreasing linearly with the logarithm of the frequency. |