IEC61010-1 关于 routing tests 的规定:
从以下的文字可看出,只有K1(接地测试), K2 和K3(耐压测试)是必须的。
有熟悉60950的专家请讲一下差别。
ANNEX K - ROUTINE TESTS (NORMATIVE) «A2
The manufacturer shall perform the tests of K.1 to K.3 on 100% of equipment produced which has both HAZARDOUS LIVE parts and ACCESSIBLE conductive parts. «A2
The ROUTINE TEST is required between supply mains connected together, and ACCESSIBLE conductive parts likely to become energized, including the protective grounding conductor. «NA
Unless it can be clearly shown that the result of tests cannot be invalidated by subsequent manufacturing stages, tests shall be made with equipment fully assembled. «A2
The equipment shall not be unwired, modified or disassembled for the test, but snap-on covers and friction- fit knobs can be removed if they would interfere with the tests. The equipment shall not be energized
during the tests, but the mains switch shall be in the on-position. «A2
Wrapping the equipment in foil is not required, nor is humidity preconditioning necessary. «A2
K.1 Protective earth «A2
A continuity test is made between the earth pin of the appliance inlet or the mains plug of plug-connected
equipment, or the PROTECTIVE CONDUCTOR TERMINAL of PERMANENTLY CONNECTED EQUIPMENT. on
the one side, and all ACCESSIBLE conductive parts which are required by 6.5.1 to be connected to the
PROTECTIVE CONDUCTOR TERMINAL on the other side. «A2
K.2 Mains circuits «A2
A test voltage as specified in annex D for BASIC INSULATION is applied between the mains TERMINALS
connected together on the one side, and all ACCESSIBLE conductive parts connected together on the other.
For this test, the contacts of any output connectors are considered as ACCESSIBLE conductive parts if, in
NORMAL USE, they could become connected to an ACCESSIBLE conductive part outside of the equipment.
«A2
The test voltage is raised to its specified value within 2 s and maintained for at least 2 s. «A2
No breakdown or repeated flashover shall occur. Corona effects and similar phenomena are disregarded.
«A2
K.3 Other circuits «A2
A test voltage is applied between the TERMINALS of floating input circuits, which can be hazardous live in
NORMAL USE, connected together on the one side, and ACCESSIBLE conductive parts connected together
on the other. «A2
A test voltage is also applied between the TERMINALS of floating output circuits, which can be
HAZARDOUS LIVE in NORMAL USE connected together on the one side, and ACCESSIBLE conductive parts
connected on the other. «A2
The value of the applied voltage in each case is 1,5 times the working voltage (see D. 1. 1). If voltage
limiting (clamping) devices operate below 1,5 times the working voltage, the value of the applied voltage
is 0,9 times the clamping voltage, but not less than that of the working voltage. «A2
NOTE - In equipment which has ACCESSIBLE conductive parts connected to the PROTECTIVE CONDUCTOR TERMINAL. a
connection ran be made to the earth pin of the appliance inlet or mains plug. During this test. the equipment should
be electrically isolated from any external earthing means. «A2
The test voltage is raised to its specified value within 2 s and maintained for at least 2 s. «A2
No breakdown or repeated flashover shall occur. Corona effects and similar phenomena are disregarded.
«A2 |
|